Dear Sung,
It seems I misread your diagram the first time, so misinterpreted
your problem.  I'm thinking now that the issue may be the cosine
correction of your sensors rather than what is going on in Radiance.
Most lux sensors have a crude calibration with respect to incident
angle.  I recommend checking your sensors against one that is more
reliable, or at least verifying that they have the proper cosine
response by measuring a point source at varying incident angles.

Best,
-Greg